Many of us (me included) have found that undetectable results of <0.006 (and low readings a bit higher) are now being reported by Labcorp as <0.014. I got no reply when I asked Labcorp about
this through their patient portal. This is from a ROCHE spec sheet (updated 28.10.2020) for their PSA assay, which Labcorp uses:
Detection Limits
LoB 0.006 ng/mL, LoD 0.010 ng/mL, LoQ 0.014 ng/mL (cobas e 411 analyzer, cobas e 601, cobas e 602 module)
LoB 0.006 ng/mL, LoD 0.014 ng/mL, LoQ 0.030 ng/mL (cobas e 801 module)
Measuring Range
0.006 - 100 ng/mL (cobas e 411 analyzer, cobas e 601, cobas e 602 modules)
0.006 - 100 ng/mL (cobas e 801 module)[Emphasis mine].
Forum brother TJ123 posted to a PSA thread last month:
TJ123 said...
Djin,
All my Labcorp PSA tests were taken at the very same clinic and the samples sent to the very same regional lab for analysis. So nothing changed except for the lower limit on my very last Labcorp test in May.
I called the Labcorp regional lab for an explanation. I was told the manufacturer changed the lower limit for their test.
Since I got no answer from Labcorp despite repeated attempts, I decided to email ROCHE, who replied yesterday:
ROCHE said...
The lower measuring range claim for the total PSA assay is 0.014 ng/mL defined by the Limit of Detection for all instruments running the test. The Limit of Blank is 0.006, which can be measured by the cobas e 411, e 601 and e 602. The Limit of Blank cannot [be] measured by the cobas e 801. It may be helpful to contact the laboratory to determine the instrument and convention used for reporting at the lower measuring limit.
[Emphasis mine]
It turns out that when you repeatedly run blanks with this test's reagents and
no PSA at all, you don't get back results of nothing or zero, but rather some number that can be as high as 0.006. This is defined as the
Limit of Blank (LoB). Clearly, you can't detect/report actual patient values in this range. The lowest value above the LoB you can
reliably distinguish from the LoB is called the
Limit of Detection (LoD).
For definitions and excellent explanations of LoB, LoD, and LoQ (Limit of Quantitation), see the text and figure in
Limit of Blank, Limit of Detection and Limit of Quantitation (2008, Full Text)
I had thought that the change to <0.014 was because uPSA tests were being run on less sensitive instruments, but I now think that is not the case; rather, someone, either at Labcorp or ROCHE, probably questioned using <0.006 and Labcorp realized (or was told) they should be using the assay's LoD (or LoQ), not the LoB, for reporting "undetectable" results. (I found a much older ROCHE spec sheet for this assay, and the numbers were the same, so nothing changed with this assay at ROCHE's end as far as I can determine.) Note that <0.014 used to be Labcorp's undetectable for uPSA until a few years ago, when they lowered it to <0.006.